When the SID is 40 inches (102 cm), the actual SID should match the displayed SID within a range of:

Prepare for the Clover Learning Radiography Image Evaluation and Quality Control Test with targeted quizzes and explanations. Master the skills needed to succeed in radiography and ensure quality control with our dynamic study materials.

Multiple Choice

When the SID is 40 inches (102 cm), the actual SID should match the displayed SID within a range of:

Explanation:
Accurate SID (source-to-image distance) indicators are essential because small deviations change magnification and exposure calculations. The tolerance for SID accuracy is typically a percentage of the indicated distance. For a 40-inch SID, the standard tolerance is ±2%. That means the actual SID should be within 0.8 inches of the displayed value (40 inches × 0.02 = 0.8 inches), which is about 2.0 cm. So the acceptable range is 39.2 to 40.8 inches (roughly 99.5 to 103 cm). This is why the correct range is ±0.8 inches (±2.0 cm).

Accurate SID (source-to-image distance) indicators are essential because small deviations change magnification and exposure calculations. The tolerance for SID accuracy is typically a percentage of the indicated distance. For a 40-inch SID, the standard tolerance is ±2%. That means the actual SID should be within 0.8 inches of the displayed value (40 inches × 0.02 = 0.8 inches), which is about 2.0 cm. So the acceptable range is 39.2 to 40.8 inches (roughly 99.5 to 103 cm). This is why the correct range is ±0.8 inches (±2.0 cm).